Hello all. Can I have your opinions about choosing the ‘right’ immersion media and lens for imaging?
The coverslip (~170um) has a RI of 1.51. The specimen can be live cells RI 1.33 in media of RI 1.33, live tissue RI 1.46, or fixed cells RI 1.33 in mounting media RI 1.44-1.47.
Can you comment on the following scenarios?
For live monolayer cells (~10um), is it better
to 1. use water lens to match the RI of media and cell
or 2. use oil to match the RI of the dish coverslip
or 3. use water dipping lens in the upright setup?
For fixed monolayer cells (~10um) in mounting media, is it better
to 1. use water lens to match the cell RI
or 2. use glycerol or silicone oil to match the RI of mounting media
or 3. use oil to match the RI of coverslip?
In the above cases, considering the sample is sitting right on the coverslip, the thickness of coverslip is dominant compared to the thickness of sample. Would it make sense to always use oil and oil lens to match coverslip, minimizing refraction?
For fixed tissues (~100um) in mounting media, is it better
to 1.use glycerol or silicone oil to match the RI of mounting media and tissue
or 2. use oil to match the RI of coverslip?
Now the tissue thickness becomes dominant than the coverslip. Would it make more sense to use glycerol lens?
For intravital multiphoton imaging directly without dorsal chamber, is it better
to 1. use water dipping lens
or 2. use glycerol dipping lens?
For intravital multiphoton imaging with a dorsal chamber, is it better
to 1. use water dipping lens
or 2. use glycerol dipping lens
or 3. use oil to match the coverslip RI?
In this case, depending on how deep I can image into the tissue, either coverslip or the sample becomes dominant. Shall I choose the immersion media based on how deep I need to image?
My apologies in advance if I missed some more important considerations or this has been discussed in the past. Please share your thoughts!