I have just small question.
I usually use the LSM980 Airyscan for imaging cells.
When I change plates to image a different one, the Airyscan detector starts rearranging its position, and I can see a change in brightness.
(rearrangement means - airy scan detector adjustment: fiber positon)
I wonder if this process affects the intensity values of the samples?
I want to compare the relative intensity values between different plates.
1 Like
It will. You need to realign the array each time you change your sample.
The coverglass bottoms of multi-well plates can be horribly unreliable with respect to thickness - and unless you are using high precision #1.5 (170um) plates you will notice quite a bit of variability in the alignment. Changing to high precision plates will help.
Thanks for your reply.
Then how can I compare the intensity at different plate.
If Airyscan detector realign affect intensity, it seems hard to comparing samples at different plate.
Do I have to use just confocal mode without airyscan?