Hi all,  
I was wondering if anybody knows of a good option for calibrating STED resolution? I’m aware of

> **[Gattaquant DNA Nanotechnologies - GATTA-STED Nanoruler](https://www.gattaquant.com/products/gatta-sted-nanoruler/gatta-sted-nanoruler.html)**

  
(and their 3D option, but not sure how that’s setup… are they all vertical maybe?)

Are there others?

I also thought I’d run some 40nm nanodiamonds as they don’t bleach!

> **[Products](http://bikanta.com/products)**

Thanks for your time.  
Neil

---

<div class="post-metadata">

**Author:** ![lightmicroscopy](https://sea2.discourse-cdn.com/flex020/user_avatar/forum.microlist.org/lightmicroscopy/32/640_2.png) [@lightmicroscopy](https://forum.microlist.org/u/lightmicroscopy)\
**Post date:** [November 3, 2020, 11:27am UTC](https://forum.microlist.org/t/sted-nanoruler-and-or-calibration/1189/2 "2020-11-03T11:27:52Z")

</div>

We are using 40-nm-beads from the [alignment sets](https://www.abberior.com/jtl-shop/abberior-Alignment-Set-) of Abberior. They can be used for up to a year or so with rather little bleaching (at least those for a 775 nm STED laser).

---

<div class="post-metadata">

**Author:** ![nranthony](https://sea2.discourse-cdn.com/flex020/user_avatar/forum.microlist.org/nranthony/32/575_2.png) [@nranthony](https://forum.microlist.org/u/nranthony)\
**Post date:** [November 17, 2020, 7:24pm UTC](https://forum.microlist.org/t/sted-nanoruler-and-or-calibration/1189/3 "2020-11-17T19:24:59Z")

</div>

Hi Thomas, thanks for the info. We just received our alignment set and we’ll proceed with that.

I found out that the bikanta nanodiamonds are are excited by the 775nm STED line. Given they don’t bleach, I wonder if they might serve as a nice option to generate STED PSFs?

Are STED donut PSFs a thing? Does anybody know?

---

<div class="post-metadata">

**Author:** ![WillGiang](https://sea2.discourse-cdn.com/flex020/user_avatar/forum.microlist.org/willgiang/32/1224_2.png) [@WillGiang](https://forum.microlist.org/u/WillGiang)\
**Post date:** [November 18, 2020, 6:04pm UTC](https://forum.microlist.org/t/sted-nanoruler-and-or-calibration/1189/4 "2020-11-18T18:04:02Z")

</div>

For STED PSFs, can’t we image the backscattered signal of gold nanoparticles?