# Offset and HyD Detectors?

**URL:** <https://forum.microlist.org/t/offset-and-hyd-detectors/1435>\
**Category:** Software\
**Tags:** confocal\
**Created:** [June 12, 2021, 2:28am UTC](https://forum.microlist.org/t/offset-and-hyd-detectors/1435 "2021-06-12T02:28:28Z")\
**Posts on this page:** 3\
**Page:** 1

<div class="post-metadata">

**Author:** ![phoTONIcs](https://avatars.discourse-cdn.com/v4/letter/p/45deac/32.png) [@phoTONIcs](https://forum.microlist.org/u/phoTONIcs)\
**Post date:** [June 12, 2021, 2:28am UTC](https://forum.microlist.org/t/offset-and-hyd-detectors/1435/1 "2021-06-12T02:28:28Z")

</div>

Probably a dumb question… but why is Smart Offset disabled for HyD detectors? If zero values are not good for PMTs, then why are they allowed for HyDs?

---

<div class="post-metadata">

**Author:** ![oQQQQQ8](https://sea2.discourse-cdn.com/flex020/user_avatar/forum.microlist.org/oqqqqq8/32/1030_2.png) [@oQQQQQ8](https://forum.microlist.org/u/oQQQQQ8)\
**Post date:** [June 17, 2021, 5:10am UTC](https://forum.microlist.org/t/offset-and-hyd-detectors/1435/2 "2021-06-17T05:10:05Z")

</div>

The bottom half of the HyD is solid state and does not leak current like a PMT circuit does. The need for PMT offset is to delete this current, so you only record signal changes due to the arrival of light. Since HyD detectors have a fixed and steady zero, no offset correction is needed.

I must say that I would have preferred that HyD detectors started at 1 instead of zero, as the sea of green can be distracting compared to a black background, when using the QLUT feature. Leica could easily do this on all-HyD systems with a simple substitution of color palette. But in mixed systems, you need to see how much zero you are getting when tweaking the PMT offsets.

---

<div class="post-metadata">

**Author:** ![phoTONIcs](https://avatars.discourse-cdn.com/v4/letter/p/45deac/32.png) [@phoTONIcs](https://forum.microlist.org/u/phoTONIcs)\
**Post date:** [June 18, 2021, 5:32pm UTC](https://forum.microlist.org/t/offset-and-hyd-detectors/1435/3 "2021-06-18T17:32:44Z")

</div>

Thanks for your reply!