# Nikon Microphot-FXA project / Hardware Parts

**URL:** <https://forum.microlist.org/t/nikon-microphot-fxa-project-hardware-parts/3060>\
**Category:** Hardware\
**Created:** [April 7, 2025, 7:40pm UTC](https://forum.microlist.org/t/nikon-microphot-fxa-project-hardware-parts/3060 "2025-04-07T19:40:29Z")\
**Posts on this page:** 13\
**Page:** 1

<div class="post-metadata">

**Author:** ![aristillius](https://sea2.discourse-cdn.com/flex020/user_avatar/forum.microlist.org/aristillius/32/1666_2.png) [@aristillius](https://forum.microlist.org/u/aristillius)\
**Post date:** [April 7, 2025, 7:40pm UTC](https://forum.microlist.org/t/nikon-microphot-fxa-project-hardware-parts/3060/1 "2025-04-07T19:40:29Z")

</div>

Continuing the discussions from [Obtaining parts and information for Nikon Microphot-FXA](https://forum.microlist.org/t/obtaining-parts-and-information-for-nikon-microphot-fxa/2333/2) and [Nikon Microphot-FXA Microscope](https://forum.microlist.org/t/nikon-microphot-fxa-microscope/2588):

Mike, (@Mdevvy)

Thanks for the status on your project. I figured I would start new conversation since the original was “closed by solution”.

I am curious if you, Barry (@MicroChemist), or anyone else out there who has one of these microscopes have found a cable wiring diagram for the FXA / FXL control/power cable? I have a PSM-4A supply that I picked up on eBay a few months ago. Very interested if the cable you have is a straight-through type (pin-to-pin) between the connectors? If that is the case, I can build one to test the power supply.

I have an interesting conundrum here. I purchased a FXA from eBay for $250 in Dec '24. The seller shipped it to me in the most unprotected cardboard box, using old present wrapping paper as cushion filler. Was very disappointed when it arrived! Especially, since I was charged $165 for shipping. The “bio hazard” label on the box may have encouraged the shipping folks to expedite the box out of their eyesight as rapidly as possible. Note how the base of the instrument is protruding through the hole in the box.

 ![Microphot-FXA_shipping_box2](https://us1.discourse-cdn.com/flex020/uploads/microlist/original/2X/d/d26a595c85dc8a1d9ebf0be9cacb616cc69511a8.jpeg)

Sure, I purchased it knowing it was a “Parts Only” unit and expected it would have its challenges. Had to start with something. Surprisingly, not much damage to the optics. The binocular assembly had a broken prism and there was some cosmetic damage to the selector buttons on the focus assembly. I found a replacement Binocular and another filter cassette because the original had some non-moving parts.

A month or two later, another FXA was posted for about the same price (different vendor, thankfully). It arrived with much better attention to packaging and shipping.

I know there are a lot of options for this model. These two have some differences between them including the base electronics interface panels. The first one shipped to me has a DB-25 connector for power input and a 4-pin circular for the Mercury Light Source Housing. But the second one is actually older and it has the SUMICON Rectangular Connector.

Any information would be greatly appreciated!

Cheers!

Troy

---

<div class="post-metadata">

**Author:** ![Mdevvy](https://avatars.discourse-cdn.com/v4/letter/m/bcef8e/32.png) [@Mdevvy](https://forum.microlist.org/u/Mdevvy)\
**Post date:** [April 10, 2025, 6:48pm UTC](https://forum.microlist.org/t/nikon-microphot-fxa-project-hardware-parts/3060/2 "2025-04-10T18:48:00Z")

</div>

So, Troy.

I like where this is going. I’d love to get mine working, like, really badly!

The PSM-1 unit and cable that I have (uselessly) are a straight through 1-1 connection through the SUMICON cable.

The 26 pins go from the PSU to the back of the scope with no funny-business.  
When turned to face each other, 1 rings continuity to 1, 2 to 2 etc.

This means we actually don’t need old school SUMICON plugs, I’d actually be happy to terminate a more readily available 26pole terminal if we can’t get SUMICONS. (I regularly make wiring looms for comms and HV components and could definitely do a neat job but would need to alter the back of our scope bodies and PSU).

Alternatively we could make a straight through cable from the SUMICONS if we can find them.

I do, however, need access to a PSM-4a 🙂

Mike.

 ![IMG_8086](https://us1.discourse-cdn.com/flex020/uploads/microlist/original/2X/6/60516a5c1ac3de901042bcda1d7e3607cc7d8d42.jpeg)

---

<div class="post-metadata">

**Author:** ![aristillius](https://sea2.discourse-cdn.com/flex020/user_avatar/forum.microlist.org/aristillius/32/1666_2.png) [@aristillius](https://forum.microlist.org/u/aristillius)\
**Post date:** [April 11, 2025, 12:25am UTC](https://forum.microlist.org/t/nikon-microphot-fxa-project-hardware-parts/3060/3 "2025-04-11T00:25:18Z")

</div>

Hey Mike,

The PSM-4a I have uses the 34-pin type of the 1600 series.

 ![PSM-4a_panel](https://us1.discourse-cdn.com/flex020/uploads/microlist/original/2X/2/207eae8c5893c57cf1b68bbf8ff784c6ce58e252.jpeg)

Going to map out the voltages across the pins to see what’s what.

I found a cache of the 1600 series connectors at [Mouser](https://www.mouser.com/c/connectors/headers-wire-housings/?m=Hirose%20Electric&series=1600). Seems they have a majority of the different pin densities available. So that is good news.

Will post my findings in a bit.

Cheers!

Troy

---

<div class="post-metadata">

**Author:** ![aristillius](https://sea2.discourse-cdn.com/flex020/user_avatar/forum.microlist.org/aristillius/32/1666_2.png) [@aristillius](https://forum.microlist.org/u/aristillius)\
**Post date:** [April 15, 2025, 2:47pm UTC](https://forum.microlist.org/t/nikon-microphot-fxa-project-hardware-parts/3060/4 "2025-04-15T14:47:32Z")

</div>

Update:

The design of the power supply and microscope are integral components to each other. Without a schematic for both, it has been an interesting challenge to map the signals between them. There are feedback circuits from the microscope that unlatch safety circuits and switch source voltages in the power supply, as well as sense circuits on the power supply side that respond to signals from the instrument. Not an uncommon design feature, when distributing multiple voltage and current sources.

I have opened both units and have been tracing the signal paths between the boards. Unfortunately, an interface cable will be necessary to complete the survey. Ordered some connectors/crimp pins to construct a harness. They will be here in the next few days.

More to follow…

Cheers!

---

<div class="post-metadata">

**Author:** ![Mdevvy](https://avatars.discourse-cdn.com/v4/letter/m/bcef8e/32.png) [@Mdevvy](https://forum.microlist.org/u/Mdevvy)\
**Post date:** [April 16, 2025, 10:19pm UTC](https://forum.microlist.org/t/nikon-microphot-fxa-project-hardware-parts/3060/5 "2025-04-16T22:19:28Z")

</div>

Hi. Would this

> **[1 PC. Hirose Sumicon 1600 set: 34 pole female, male connector, stopper, cover ...](https://www.ebay.co.uk/itm/326271693174?mkcid=16&mkevt=1&mkrid=711-127632-2357-0&ssspo=4_8Ih9ZST-a&sssrc=4429486&ssuid=hsbqqHkjQ_-&var=&widget_ver=artemis&media=COPY)**
>
> 1 PC.Hirose Sumicon 1600 set: 34 pole female connector, male connector, stopper, cover case. 1 piece Hirose Sumicon 1600 cover case 34 pole side lock type P-1634A-C(50) CL0216-0034-4-50. 1 piece Hirose Sumicon 1600 female 34 pole connector...

Be what we’re after plus some conductors and sheathing?

Also do you require a male one end, female the other, or both matching ends if the scope and PSU are matching the opposite.

M.

---

<div class="post-metadata">

**Author:** ![aristillius](https://sea2.discourse-cdn.com/flex020/user_avatar/forum.microlist.org/aristillius/32/1666_2.png) [@aristillius](https://forum.microlist.org/u/aristillius)\
**Post date:** [April 17, 2025, 2:09am UTC](https://forum.microlist.org/t/nikon-microphot-fxa-project-hardware-parts/3060/6 "2025-04-17T02:09:02Z")

</div>

Mike,

Yes, that is an excellent find with both M and F terminations and two shells. The scope does use the (M) connector and an (F) on the PSM-4a. So the kit you have found seems to be a good choice. The part numbers align fairly well with what I ordered. The only major difference is the connector shell type. The ones I ordered below have a CA(50) which is different in how it latches onto the stopper bracket. You may need to get a second stopper bracket to complement the kit. I would also recommend looking for a kit that contains the CA(50) variant.

I hesitated to post the part numbers until they arrived (today) to ensure I had ordered the correct ones.

P-1634A-CA(50): Connector Shells (Top Touch Lock Type)  
PC-1634(10): (M) 34-pin insert  
SC-1634(10): (F) 34-pin insert  
PC-1600-112: (M) crimp pins  
SC-1600-112; (F) crimp sockets

I am using crimp pins, so I can reconfigure pin assignments if necessary. I have 24AWG multistrand wire and sheathing mesh from another project. Will be assembling the harness over the next two evenings, since my extended holiday is over and I return to work in the morning. Should have an updated wire diagram by Sunday.

Have you opened the PSM-1 unit, and looked at the board layout? Could you tell me the board manufacturer and nomenclature, please? The PSM-4a uses a SANDEN S-PS-0216 board and utilizes 33 of the 34 pins on the Control Connector. If the PSM-1 uses same board, there is a chance it could be converted to emulate the PSM-4a. I have been keeping an eye out for another PSM-4a, unfortunately, they are becoming scarce.

Honestly, I am not confident the one I have is functional, yet. The signals and source voltages are not what I had expected. Although, I suspect it might be due to the unbalanced circuits without being integrated with the scope.

Cheers!

Troy

---

<div class="post-metadata">

**Author:** ![aristillius](https://sea2.discourse-cdn.com/flex020/user_avatar/forum.microlist.org/aristillius/32/1666_2.png) [@aristillius](https://forum.microlist.org/u/aristillius)\
**Post date:** [April 17, 2025, 2:24am UTC](https://forum.microlist.org/t/nikon-microphot-fxa-project-hardware-parts/3060/7 "2025-04-17T02:24:48Z")

</div>

I found the kit that contains the top touch lock:

> **[1 PC. Hirose Sumicon 1600 set: 34 pole fem., male connector, stopper, top touch...](https://www.ebay.co.uk/itm/326282888428)**
>
> 1 PC.Hirose Sumicon 1600 set: 34 pole female connector, male connector, stopper,top touch lock case. 1 piece Hirose Sumicon 1600 cover case 34 pole top touch CL0216-0111-3-50 P-1643A-CA(50). 1 piece Hirose Sumicon 1600 female 34 pole connector...

---

<div class="post-metadata">

**Author:** ![Mdevvy](https://avatars.discourse-cdn.com/v4/letter/m/bcef8e/32.png) [@Mdevvy](https://forum.microlist.org/u/Mdevvy)\
**Post date:** [April 19, 2025, 6:02pm UTC](https://forum.microlist.org/t/nikon-microphot-fxa-project-hardware-parts/3060/8 "2025-04-19T18:02:39Z")

</div>

Hi Troy.

Images attached of my PSA-1 with the top off.

 ![IMG_8174](https://us1.discourse-cdn.com/flex020/uploads/microlist/original/2X/7/705b81d1817a09cea594b327e8cffbd200555e93.jpeg)

Did you manage to get yours working?

Can you tell me if it’s the PSA-1 you actually need.

FYI, the board inside looks like a very early generation.

Says NIKON on it as if it’s the manufacturer with ‘made in Japan’ alongside. Underside of the board boasts the index ‘PCB 1’.

All pins look to be in use on this with some jumpers going to other places in the box, transformers and the like.

Hope this is helpful!  
I would suggest a meetup at the some point so we could plug each other in, although not sure of the rules of engagement here. I’m Oxford based.

Mike.

---

<div class="post-metadata">

**Author:** ![aristillius](https://sea2.discourse-cdn.com/flex020/user_avatar/forum.microlist.org/aristillius/32/1666_2.png) [@aristillius](https://forum.microlist.org/u/aristillius)\
**Post date:** [June 19, 2025, 4:04am UTC](https://forum.microlist.org/t/nikon-microphot-fxa-project-hardware-parts/3060/9 "2025-06-19T04:04:07Z")

</div>

Mike,

I hope this message finds you doing well. After my last post I was involved in a vehicle accident and it has taken me a few months with recovery. I am now moving about and have made some progress on the microscope.

The pin-to-pin assignment works! 1-to-1, 2-to-2, etc. I have been attempting to map the signals between the PSM-4a and the FXA. A few are straight forward but others have intermittent signals or variable voltages. It has been challenging to map them without schematics.

Unfortunately, I am not in the U.K. And as awesome as it would be for us schedule a time to meet, it would require some logistics that I currently wouldn’t be able to accomodate. 🙂

I am curious if you have had any progress on your end.

Cheers!

Troy

---

<div class="post-metadata">

**Author:** ![Mdevvy](https://avatars.discourse-cdn.com/v4/letter/m/bcef8e/32.png) [@Mdevvy](https://forum.microlist.org/u/Mdevvy)\
**Post date:** [March 16, 2026, 6:26pm UTC](https://forum.microlist.org/t/nikon-microphot-fxa-project-hardware-parts/3060/10 "2026-03-16T18:26:50Z")

</div>

Hi Troy.

Just warming this thread up as I now have a PSM-4 on the way (thanks to an eBay alert I set up a year ago) and have had delivered the hirose connector and pin kit.

Off to make the loom now and when I plug everything in, maybe I’ll have something fun to report. Wishing you good health.

Mike.

---

<div class="post-metadata">

**Author:** ![aristillius](https://sea2.discourse-cdn.com/flex020/user_avatar/forum.microlist.org/aristillius/32/1666_2.png) [@aristillius](https://forum.microlist.org/u/aristillius)\
**Post date:** [March 20, 2026, 3:04am UTC](https://forum.microlist.org/t/nikon-microphot-fxa-project-hardware-parts/3060/11 "2026-03-20T03:04:32Z")

</div>

Hi Mike,

It is good to hear from you! I am doing well, thank you. Hope you are too!

That is great news on finding a PSM-4. They are becoming a very scarce species for sure. The pin sequence for the PSM-4a is: Pin 1 to Pin 1, Pin 2 to Pin 2, etc., using the 34pin Sumicons. I am not sure if the PSM-4 has similar configuration.

 ![20250329_185613](https://us1.discourse-cdn.com/flex020/uploads/microlist/original/2X/1/12c200a12f101f58fd25f9727bd25d9756e9bb27.jpeg)

The two Microphot FXA’s side by side. The first one (left) has the 34 pin Sumicon and the second unit (right) has a DB-15 for DC input power. I have not worked with the DB-15 instrument, yet. It has been a donor for parts.

Was successful in building a harness for the first unit! Both power supply and instrument seem to work. The microprocessor booted up and the LCD display and menu selections work. I can configure most of the parameters for selecting optical paths and other functions, but it did have some problems. These instruments are configured to integrate specific 35mm film camera / dark boxes, but I have a Nikon D7100 camera mounted on the scope. Most of the programmable camera settings on the instrument are unusable for my setup.

A few internal optic issues and an incorrectly figured objective have limited my imaging / observations so far. I am very excited with the results and have started working on setting up the instrument for mineralogical applications. Below is my first attempt at focus-stacking, using 1µm focus stops across 121 µm of field depth.

 ![Rutile_acicular_Sup7(B,R50,S1)](https://us1.discourse-cdn.com/flex020/uploads/microlist/original/2X/8/887ea88cc79c561fd8776a1e1a6ecfea89983e72.jpeg)

```auto
Hematite and Rutile inclusions in a host quartz crystal; approx. 1mm FOV at ~62x magnification.

```

Keep me posted on your progress! Very much looking forward to hearing good news, sir! 🙂

Cheers!

-Troy

---

<div class="post-metadata">

**Author:** ![Mdevvy](https://avatars.discourse-cdn.com/v4/letter/m/bcef8e/32.png) [@Mdevvy](https://forum.microlist.org/u/Mdevvy)\
**Post date:** [March 30, 2026, 5:59pm UTC](https://forum.microlist.org/t/nikon-microphot-fxa-project-hardware-parts/3060/12 "2026-03-30T17:59:56Z")

</div>

Hi.

PSM-4a delivered.

Had a blown 4a fuse.

Wants a 2a fuse for my 240v setup. Flicked the switch on the front to change from 110 to 240 and replaced with the, now required, 2a fuse.

Blows it every time after switching it on at the front. Haven’t found the culprit yet. But loom made and microscope waiting…

I’ll update on any progress!

---

<div class="post-metadata">

**Author:** ![MicroChemist](https://sea2.discourse-cdn.com/flex020/user_avatar/forum.microlist.org/microchemist/32/1964_2.png) [@MicroChemist](https://forum.microlist.org/u/MicroChemist)\
**Post date:** [June 16, 2026, 3:09am UTC](https://forum.microlist.org/t/nikon-microphot-fxa-project-hardware-parts/3060/13 "2026-06-16T03:09:06Z")

</div>

Hi Troy, I picked-up PSM-4a power supply and I was nervous, since I had no wiring diagrams, to do anything, but as chance would have it I found another PSM-4A on the cheap and while looking at the connector it looked like the female to the microscopes male! So I took the female connector out of the supply that was oldest looking and drew a wire guide and then cut the connector out. Then taking a chance that nikon was not “Apple”, I wired 2 foot wires from the pins in the connector back to the wiring diagram locations I had mapped out inside the supply. I plugged the thing into the microscope and Turned ON the POWER. Yes it worked just fine ! So they are a One to One with no crossed wires ! I carefully sealed all the wires together with Hot Glue, as well as the opening they pass through in the supply to keep everything insulated. i used 22AWG stranded wire, silicone insulated. This is a super flexible wire with high strand count that can easily handle 200deg C, 3000v, 1amp according to manufacturer. The entire supply can’t deliver over 4A at 110 v. Having no schematic, the only thing I can say is that there are15 different supply voltages, so most of the individual supplies are just milli-anps. One thing to keep in mind is that the individual supplies are not using shielded wire, so there is a case ground on the back of the supply and the microscope should be grouded to it .