# Motorized Stage quits during time course

**URL:** https://forum.microlist.org/t/motorized-stage-quits-during-time-course/1232
**Category:** Hardware
**Tags:** micro-manager
**Created:** [December 10, 2020, 3:48pm UTC](https://forum.microlist.org/t/motorized-stage-quits-during-time-course/1232 "2020-12-10T15:48:49Z")
**Posts on this page:** 2
**Page:** 1

<div class="post-metadata">

### Author: ![Heather\_BrownHarding](https://sea2.discourse-cdn.com/flex020/user_avatar/forum.microlist.org/heather_brownharding/32/685_2.png) [@Heather\_BrownHarding](https://forum.microlist.org/u/Heather_BrownHarding)
#### Post date: [December 10, 2020, 3:48pm UTC](https://forum.microlist.org/t/motorized-stage-quits-during-time-course/1232/1 "2020-12-10T15:48:50Z")

</div>

We have acquired a used Olympus IX81 with a Sigma Koki XY motorized stage. We are doing time courses with multi ROI, and after ~4 time points the stage becomes unresponsive. We are running with micro-manager 2.0. Has anyone encountered a stage suddenly becoming unresponsive to a computer but still ok with the joystick?

Here is the config file if it is of any use–  
# Generated by Configurator on Mon Nov 30 14:06:55 EST 2020

```
# Reset
Property,Core,Initialize,0

# Devices
Device,COM1,SerialManager,COM1
Device,COM6,SerialManager,COM6
Device,OlympusHub,Olympus,OlympusHub
Device,Objective,Olympus,Objective
Device,LightPath,Olympus,LightPath
Device,Shutter1,Olympus,Shutter1
Device,Shutter2,Olympus,Shutter2
Device,TransmittedLamp,Olympus,TransmittedLamp
Device,FluorescenceLamp,Olympus,FluorescenceLamp
Device,ManualFocus,Olympus,ManualFocus
Device,AutoFocusZDC,Olympus,AutoFocusZDC
Device,AutoFocus Stage,Utilities,AutoFocus Stage
Device,TUCam,Tucsen_x64,TUCam
Device,XYStage,SigmaKoki,XYStage

# Pre-init settings for devices
Property,OlympusHub,Port,COM1
Property,XYStage,Move per step(micron),2.0000
Property,XYStage,Port,COM6

# Pre-init settings for COM ports
Property,COM1,AnswerTimeout,500.0000
Property,COM1,BaudRate,19200
Property,COM1,DataBits,8
Property,COM1,DelayBetweenCharsMs,0.0000
Property,COM1,Fast USB to Serial,Disable
Property,COM1,Handshaking,Off
Property,COM1,Parity,Even
Property,COM1,StopBits,2
Property,COM1,Verbose,1
Property,COM6,AnswerTimeout,500.0000
Property,COM6,BaudRate,9600
Property,COM6,DTR,Disable
Property,COM6,DataBits,8
Property,COM6,DelayBetweenCharsMs,0.0000
Property,COM6,Fast USB to Serial,Disable
Property,COM6,Handshaking,Off
Property,COM6,Parity,None
Property,COM6,StopBits,1
Property,COM6,Verbose,1

# Hub (parent) references

# Initialize
Property,Core,Initialize,1

# Delays

# Focus directions
FocusDirection,ManualFocus,0
FocusDirection,AutoFocus Stage,0

# Roles
Property,Core,Camera,TUCam
Property,Core,Focus,ManualFocus
Property,Core,AutoShutter,1

# Camera-synchronized devices

# Labels
# Objective
Label,Objective,5,Position-6
Label,Objective,4,Position-5
Label,Objective,3,Position-4
Label,Objective,2,Position-3
Label,Objective,1,Position-2
Label,Objective,0,Position-1
# LightPath
Label,LightPath,1,Side Port
Label,LightPath,0,Eye
# TransmittedLamp
Label,TransmittedLamp,1,On
Label,TransmittedLamp,0,Off
# FluorescenceLamp
Label,FluorescenceLamp,1,On
Label,FluorescenceLamp,0,Off

# Configuration presets
# Group: Objective type ZDC
# Preset: UPlanSApo60XO
ConfigGroup,Objective type ZDC,UPlanSApo60XO,AutoFocusZDC,ObjectiveTypeSetting,UPlanSApo60XO

# Preset: LCPlanFl 20X
ConfigGroup,Objective type ZDC,LCPlanFl 20X,AutoFocusZDC,ObjectiveTypeSetting,LCPlanFl 20X

# Preset: LUCPlanFlN20X
ConfigGroup,Objective type ZDC,LUCPlanFlN20X,AutoFocusZDC,ObjectiveTypeSetting,LUCPlanFlN20X

# Preset: UPlanFlN60XOI
ConfigGroup,Objective type ZDC,UPlanFlN60XOI,AutoFocusZDC,ObjectiveTypeSetting,UPlanFlN60XOI

# Preset: UPlanFlN100XO
ConfigGroup,Objective type ZDC,UPlanFlN100XO,AutoFocusZDC,ObjectiveTypeSetting,UPlanFlN100XO

# Preset: UPlanFlN100XOI
ConfigGroup,Objective type ZDC,UPlanFlN100XOI,AutoFocusZDC,ObjectiveTypeSetting,UPlanFlN100XOI

# Preset: PlanApo 60X O-LSM
ConfigGroup,Objective type ZDC,PlanApo 60X O-LSM,AutoFocusZDC,ObjectiveTypeSetting,PlanApo 60X O/LSM

# Preset: PlanApo 60XO3
ConfigGroup,Objective type ZDC,PlanApo 60XO3,AutoFocusZDC,ObjectiveTypeSetting,PlanApo 60XO3

# Preset: UPlanFl 100XO3
ConfigGroup,Objective type ZDC,UPlanFl 100XO3,AutoFocusZDC,ObjectiveTypeSetting,UPlanFl 100XO3

# Preset: UPLFLN60X
ConfigGroup,Objective type ZDC,UPLFLN60X,AutoFocusZDC,ObjectiveTypeSetting,UPLFLN60X

# Preset: UPlanSApo40X
ConfigGroup,Objective type ZDC,UPlanSApo40X,AutoFocusZDC,ObjectiveTypeSetting,UPlanSApo40X

# Preset: PlanApo 60X O TIRFM-SP
ConfigGroup,Objective type ZDC,PlanApo 60X O TIRFM-SP,AutoFocusZDC,ObjectiveTypeSetting,PlanApo 60X O TIRFM-SP

# Preset: None
ConfigGroup,Objective type ZDC,None,AutoFocusZDC,ObjectiveTypeSetting,None

# Preset: UplanFlN40X
ConfigGroup,Objective type ZDC,UplanFlN40X,AutoFocusZDC,ObjectiveTypeSetting,UplanFlN40X

# Preset: UPlanFlN60X
ConfigGroup,Objective type ZDC,UPlanFlN60X,AutoFocusZDC,ObjectiveTypeSetting,UPlanFlN60X

# Preset: UApo 40XOI3 -340
ConfigGroup,Objective type ZDC,UApo 40XOI3 -340,AutoFocusZDC,ObjectiveTypeSetting,UApo 40XOI3 /340

# Preset: UPlanSApo20X
ConfigGroup,Objective type ZDC,UPlanSApo20X,AutoFocusZDC,ObjectiveTypeSetting,UPlanSApo20X

# Preset: UPlanFl 40X O-SP
ConfigGroup,Objective type ZDC,UPlanFl 40X O-SP,AutoFocusZDC,ObjectiveTypeSetting,UPlanFl 40X O-SP

# Preset: UApo40XW
ConfigGroup,Objective type ZDC,UApo40XW,AutoFocusZDC,ObjectiveTypeSetting,UApo40XW

# Preset: PlanApo40XW-LSM
ConfigGroup,Objective type ZDC,PlanApo40XW-LSM,AutoFocusZDC,ObjectiveTypeSetting,PlanApo40XW/LSM

# Preset: UPLSAPO60XO
ConfigGroup,Objective type ZDC,UPLSAPO60XO,AutoFocusZDC,ObjectiveTypeSetting,UPLSAPO60XO

# Preset: UPlanApo 40X OI3
ConfigGroup,Objective type ZDC,UPlanApo 40X OI3,AutoFocusZDC,ObjectiveTypeSetting,UPlanApo 40X OI3

# Preset: LUCPlanFlN60X
ConfigGroup,Objective type ZDC,LUCPlanFlN60X,AutoFocusZDC,ObjectiveTypeSetting,LUCPlanFlN60X

# Preset: LCPlanFl 60X
ConfigGroup,Objective type ZDC,LCPlanFl 60X,AutoFocusZDC,ObjectiveTypeSetting,LCPlanFl 60X

# Preset: LUCPLFLN40X
ConfigGroup,Objective type ZDC,LUCPLFLN40X,AutoFocusZDC,ObjectiveTypeSetting,LUCPLFLN40X

# Preset: UAPO40XO Iris3-340
ConfigGroup,Objective type ZDC,UAPO40XO Iris3-340,AutoFocusZDC,ObjectiveTypeSetting,UAPO40XO Iris3/340

# Preset: UPLFLN100XOI
ConfigGroup,Objective type ZDC,UPLFLN100XOI,AutoFocusZDC,ObjectiveTypeSetting,UPLFLN100XOI

# Preset: PlanApoN60XO
ConfigGroup,Objective type ZDC,PlanApoN60XO,AutoFocusZDC,ObjectiveTypeSetting,PlanApoN60XO

# Preset: UApoN100XOTIRF
ConfigGroup,Objective type ZDC,UApoN100XOTIRF,AutoFocusZDC,ObjectiveTypeSetting,UApoN100XOTIRF

# Preset: PlanApo 100XO3
ConfigGroup,Objective type ZDC,PlanApo 100XO3,AutoFocusZDC,ObjectiveTypeSetting,PlanApo 100XO3

# Preset: UPlanApo60XW
ConfigGroup,Objective type ZDC,UPlanApo60XW,AutoFocusZDC,ObjectiveTypeSetting,UPlanApo60XW

# Preset: UPlanAPO 20X
ConfigGroup,Objective type ZDC,UPlanAPO 20X,AutoFocusZDC,ObjectiveTypeSetting,UPlanAPO 20X

# Preset: UPlanSApo100XO
ConfigGroup,Objective type ZDC,UPlanSApo100XO,AutoFocusZDC,ObjectiveTypeSetting,UPlanSApo100XO

# Preset: UPlanFl40X
ConfigGroup,Objective type ZDC,UPlanFl40X,AutoFocusZDC,ObjectiveTypeSetting,UPlanFl40X

# Preset: LUCPlanFlN40X
ConfigGroup,Objective type ZDC,LUCPlanFlN40X,AutoFocusZDC,ObjectiveTypeSetting,LUCPlanFlN40X

# Preset: LCPlanFl 40X
ConfigGroup,Objective type ZDC,LCPlanFl 40X,AutoFocusZDC,ObjectiveTypeSetting,LCPlanFl 40X

# Preset: UPlanAPo 100XOI3
ConfigGroup,Objective type ZDC,UPlanAPo 100XOI3,AutoFocusZDC,ObjectiveTypeSetting,UPlanAPo 100XOI3

# Preset: LUCPLFLN 20X
ConfigGroup,Objective type ZDC,LUCPLFLN 20X,AutoFocusZDC,ObjectiveTypeSetting,LUCPLFLN 20X

# Preset: UApo 40X3 -340
ConfigGroup,Objective type ZDC,UApo 40X3 -340,AutoFocusZDC,ObjectiveTypeSetting,UApo 40X3 /340

# Preset: UPlanFlN20X
ConfigGroup,Objective type ZDC,UPlanFlN20X,AutoFocusZDC,ObjectiveTypeSetting,UPlanFlN20X

# Preset: UPLFLN100XO2
ConfigGroup,Objective type ZDC,UPLFLN100XO2,AutoFocusZDC,ObjectiveTypeSetting,UPLFLN100XO2

# Preset: UPlanApo60XW-IR
ConfigGroup,Objective type ZDC,UPlanApo60XW-IR,AutoFocusZDC,ObjectiveTypeSetting,UPlanApo60XW/IR

# Preset: ApoN60XOTIRF
ConfigGroup,Objective type ZDC,ApoN60XOTIRF,AutoFocusZDC,ObjectiveTypeSetting,ApoN60XOTIRF

# Preset: LUCPlanFl 40X
ConfigGroup,Objective type ZDC,LUCPlanFl 40X,AutoFocusZDC,ObjectiveTypeSetting,LUCPlanFl 40X

# Preset: UPlanFlN40XO
ConfigGroup,Objective type ZDC,UPlanFlN40XO,AutoFocusZDC,ObjectiveTypeSetting,UPlanFlN40XO

# Preset: UPlanFl 20X
ConfigGroup,Objective type ZDC,UPlanFl 20X,AutoFocusZDC,ObjectiveTypeSetting,UPlanFl 20X

# Preset: UPlanSApo60XW
ConfigGroup,Objective type ZDC,UPlanSApo60XW,AutoFocusZDC,ObjectiveTypeSetting,UPlanSApo60XW

# Preset: UApo 20X3 -340
ConfigGroup,Objective type ZDC,UApo 20X3 -340,AutoFocusZDC,ObjectiveTypeSetting,UApo 20X3 /340

# Preset: UPLanFl 60XOI3
ConfigGroup,Objective type ZDC,UPLanFl 60XOI3,AutoFocusZDC,ObjectiveTypeSetting,UPLanFl 60XOI3

# Preset: UPLSAPO40X2
ConfigGroup,Objective type ZDC,UPLSAPO40X2,AutoFocusZDC,ObjectiveTypeSetting,UPLSAPO40X2

# Preset: SLCPlanFl 40X
ConfigGroup,Objective type ZDC,SLCPlanFl 40X,AutoFocusZDC,ObjectiveTypeSetting,SLCPlanFl 40X

# Preset: UPlanApo 60X
ConfigGroup,Objective type ZDC,UPlanApo 60X,AutoFocusZDC,ObjectiveTypeSetting,UPlanApo 60X

# Group: light
# Preset: 1
ConfigGroup,light,1,TransmittedLamp,ClosedPosition,1

# Preset: 0
ConfigGroup,light,0,TransmittedLamp,ClosedPosition,0

# Group: Measure off set ZDC
# Preset: Off
ConfigGroup,Measure off set ZDC,Off,AutoFocusZDC,MeasureOffset,Off

# Preset: Now
ConfigGroup,Measure off set ZDC,Now,AutoFocusZDC,MeasureOffset,Now

# Group: Channel

# Group: ZDC search range 
# Preset: NewPreset
ConfigGroup,ZDC search range ,NewPreset,AutoFocusZDC,SearchRange,0.0000

# Group: System
# Preset: Startup

# Group: control
# Preset: Computer
ConfigGroup,control,Computer,OlympusHub,Control,Computer

# Preset: Manual + Computer
ConfigGroup,control,Manual + Computer,OlympusHub,Control,Manual + Computer

# PixelSize settings
```

---

<div class="post-metadata">

### Author: ![nico.stuurman](https://sea2.discourse-cdn.com/flex020/user_avatar/forum.microlist.org/nico.stuurman/32/302_2.png) [@nico.stuurman](https://forum.microlist.org/u/nico.stuurman)
#### Post date: [December 10, 2020, 4:07pm UTC](https://forum.microlist.org/t/motorized-stage-quits-during-time-course/1232/2 "2020-12-10T16:07:42Z")

</div>

Welcome here @Heather_BrownHarding!

FWIW, we have a Sigma Koki controller as well, that has been working for a long time without problems, but I have vague memories of initial trouble (and of course forgot how we worked around that). Maybe it was something along the lines of never send another command to move if the previous move did not complete yet. It may be helpful to generate a Trouble Report (Help \> Report a Problem). That allows me to look at a lot of debug output that maybe explains what is going on.