# Live cell heat shock on microscope

**URL:** <https://forum.microlist.org/t/live-cell-heat-shock-on-microscope/720>\
**Category:** Hardware\
**Tags:** confocal\
**Created:** [November 20, 2019, 3:12pm UTC](https://forum.microlist.org/t/live-cell-heat-shock-on-microscope/720 "2019-11-20T15:12:12Z")\
**Posts on this page:** 13\
**Page:** 1

<div class="post-metadata">

**Author:** ![Chaowei](https://avatars.discourse-cdn.com/v4/letter/c/db5fbb/32.png) [@Chaowei](https://forum.microlist.org/u/Chaowei)\
**Post date:** [November 20, 2019, 3:12pm UTC](https://forum.microlist.org/t/live-cell-heat-shock-on-microscope/720/1 "2019-11-20T15:12:12Z")

</div>

Hi,  
I’m looking for suggestions on how to induce heat shock in live cells/yeast on a microscope. What device do you use to meet the fast temperature change in order to induce heat shock? I found the VAHEAT system from BostonElectronics. Has anyone used it?

Thank you!😁

---

<div class="post-metadata">

**Author:** ![Josef\_Gotzmann](https://sea2.discourse-cdn.com/flex020/user_avatar/forum.microlist.org/josef_gotzmann/32/241_2.png) [@Josef\_Gotzmann](https://forum.microlist.org/u/Josef_Gotzmann)\
**Post date:** [November 25, 2019, 9:07am UTC](https://forum.microlist.org/t/live-cell-heat-shock-on-microscope/720/2 "2019-11-25T09:07:16Z")

</div>

You could use the Cherry-Temp - allows for rather quick temperature ramping. We use one of those.  
NO COMMERCIAL INTEREST

> **[Heater Cooler for Microscopy | Cherry Biotech](https://www.cherrybiotech.com/heater-cooler-for-microscopy)**
>
> We developed the world's fastest heater cooler for microscopy. Originally designed for ultra-fast qPCR military applications, we improved this temperature control plateform to perfectly fit live-cell imaging experimentations. CherryTemp enables...

  
Best  
Josef

---

<div class="post-metadata">

**Author:** ![Chaowei](https://avatars.discourse-cdn.com/v4/letter/c/db5fbb/32.png) [@Chaowei](https://forum.microlist.org/u/Chaowei)\
**Post date:** [November 25, 2019, 3:08pm UTC](https://forum.microlist.org/t/live-cell-heat-shock-on-microscope/720/3 "2019-11-25T15:08:16Z")

</div>

Thank you Josef. I asked for the price, it’s too expensive… Seems like not many people do heat shock on a microscope…

---

<div class="post-metadata">

**Author:** ![vrkrishn](https://sea2.discourse-cdn.com/flex020/user_avatar/forum.microlist.org/vrkrishn/32/414_2.png) [@vrkrishn](https://forum.microlist.org/u/vrkrishn)\
**Post date:** [December 11, 2019, 4:58am UTC](https://forum.microlist.org/t/live-cell-heat-shock-on-microscope/720/4 "2019-12-11T04:58:10Z")

</div>

To induce a heat shock, I place the sample in a plate incubator for 5 mins and transfer the sample to the microscope with an on-stage incubator maintained at the temperature of interest.  
[http://www.oko-lab.com/live-cell-imaging/stage-top-digital-gas/28-configrator-article/196-top-satge-incubator-3](http://www.oko-lab.com/live-cell-imaging/stage-top-digital-gas/28-configrator-article/196-top-satge-incubator-3)

---

<div class="post-metadata">

**Author:** ![Chaowei](https://avatars.discourse-cdn.com/v4/letter/c/db5fbb/32.png) [@Chaowei](https://forum.microlist.org/u/Chaowei)\
**Post date:** [December 11, 2019, 10:02pm UTC](https://forum.microlist.org/t/live-cell-heat-shock-on-microscope/720/5 "2019-12-11T22:02:38Z")

</div>

Thank you vrkrishn! What objective do you use? Our user needs the 100x for yeast. I’m not sure how the higher temperature affects the objective.

---

<div class="post-metadata">

**Author:** ![vrkrishn](https://sea2.discourse-cdn.com/flex020/user_avatar/forum.microlist.org/vrkrishn/32/414_2.png) [@vrkrishn](https://forum.microlist.org/u/vrkrishn)\
**Post date:** [December 11, 2019, 10:25pm UTC](https://forum.microlist.org/t/live-cell-heat-shock-on-microscope/720/6 "2019-12-11T22:25:27Z")

</div>

we use Olympus TIRF UApo N100X N.A. = 1.49 oil immersion objective,

> **[Olympus UAPON 100x Oil TIRF Microscope Objective](https://microscopecentral.com/products/olympus-uapon-100x-oil-tirf-microscope-objective)**
>
> Olympus UAPON 100x Oil Microscope Objective for BX, IX, CX series microscopes. RMS Threads N.A. 1.49 W.D. 0.1 Part Number 1-U2B725 Brand New

you can rotate the dial and use the for higher temperature it works fine at least upto 42 C.

---

<div class="post-metadata">

**Author:** ![Chaowei](https://avatars.discourse-cdn.com/v4/letter/c/db5fbb/32.png) [@Chaowei](https://forum.microlist.org/u/Chaowei)\
**Post date:** [December 11, 2019, 10:39pm UTC](https://forum.microlist.org/t/live-cell-heat-shock-on-microscope/720/7 "2019-12-11T22:39:18Z")

</div>

That’s good to know. I’m always a little worried about objective performance and lifetime under higher temperature, but there doesn’t seem to have any ways to avoid it.

---

<div class="post-metadata">

**Author:** ![talley](https://sea2.discourse-cdn.com/flex020/user_avatar/forum.microlist.org/talley/32/3_2.png) [@talley](https://forum.microlist.org/u/talley)\
**Post date:** [December 19, 2019, 6:04pm UTC](https://forum.microlist.org/t/live-cell-heat-shock-on-microscope/720/8 "2019-12-19T18:04:42Z")

</div>

My understanding is that repeated changes in temperature are more detrimental than simply heating a lens to something like 37. We tend to keep lenses on “live cell” rigs permanently heated to avoid unnecessary temperature cycles.

(Though if you’re pushing it much higher than 37, I would definitely check with the manufacturer)

---

<div class="post-metadata">

**Author:** ![Chaowei](https://avatars.discourse-cdn.com/v4/letter/c/db5fbb/32.png) [@Chaowei](https://forum.microlist.org/u/Chaowei)\
**Post date:** [December 19, 2019, 6:44pm UTC](https://forum.microlist.org/t/live-cell-heat-shock-on-microscope/720/9 "2019-12-19T18:44:46Z")

</div>

Thanks Telly! We will probably just purchase another objective just for the heatshock users.

---

<div class="post-metadata">

**Author:** ![Jaroslav](https://sea2.discourse-cdn.com/flex020/user_avatar/forum.microlist.org/jaroslav/32/751_2.png) [@Jaroslav](https://forum.microlist.org/u/Jaroslav)\
**Post date:** [April 1, 2021, 8:58am UTC](https://forum.microlist.org/t/live-cell-heat-shock-on-microscope/720/10 "2021-04-01T08:58:18Z")

</div>

Hi, did you try the VAHEAT device in the end? Because I think it is a good fit for your application. You can program also all sorts of complex heat shock patterns with it. Just a clarification, this device is a product of Interherence a German startup, Boston Electronics is the US distributor. Check out the new development around VAHEAT device here: [VAHEAT - Micro heating system](https://interherence.com/vaheat/)

Thanks to the design of the device, there is little heat exchange to the objective. You heat up just a small volume of the sample in your field of view. We measured the PSF of a 100x immersion objective and the PSF stayed the same up until 80°C.

Full disclosure, I have just started working for Interherence now. I encourage you to judge it for yourself on the web or request an on site demo through the company website or Boston Electronics. Among the 50 or so devices we sold since last year, there is also a customer working with yeast.

---

<div class="post-metadata">

**Author:** ![Chaowei](https://avatars.discourse-cdn.com/v4/letter/c/db5fbb/32.png) [@Chaowei](https://forum.microlist.org/u/Chaowei)\
**Post date:** [April 9, 2021, 3:57pm UTC](https://forum.microlist.org/t/live-cell-heat-shock-on-microscope/720/11 "2021-04-09T15:57:22Z")

</div>

Hi Jaroslav,

Yes, we did buy it. It is user friendly to both mammalian cells and yeast. Looking forward to receiving and testing it!  
No commercial Interest

---

<div class="post-metadata">

**Author:** ![Jaroslav](https://sea2.discourse-cdn.com/flex020/user_avatar/forum.microlist.org/jaroslav/32/751_2.png) [@Jaroslav](https://forum.microlist.org/u/Jaroslav)\
**Post date:** [April 10, 2021, 9:19pm UTC](https://forum.microlist.org/t/live-cell-heat-shock-on-microscope/720/12 "2021-04-10T21:19:58Z")

</div>

Great, I would be very interested in your feedback about the performance in heat shocks. This is one of the applications the device was specifically designed for. Also, if you need anything related to vaheat, just contact me at: [jaroslav@interherence.com](mailto:jaroslav@interherence.com)

---

<div class="post-metadata">

**Author:** ![semliki16](https://avatars.discourse-cdn.com/v4/letter/s/ee7513/32.png) [@semliki16](https://forum.microlist.org/u/semliki16)\
**Post date:** [September 11, 2023, 9:23pm UTC](https://forum.microlist.org/t/live-cell-heat-shock-on-microscope/720/13 "2023-09-11T21:23:40Z")

</div>

HI Chaowei,  
I can recommend our device CellEraser - it can heat cells to any desired temperature for any needed time. The heating is local i.e. heated spot might be 500-1000 mkm in diameter and other cells around will not be heated (serve as a control). The heating is internal via millimeter wave radiation (plastic, glass and other materials are not heated ). Temperature calibration is precise (via small thermocouple or melting of film of long-chain alcohols with known Tm).  
See our website: [www.celleraser.com](http://www.celleraser.com)  
Sincerely,  
Andy