# Issue with µmanager for IX83

**URL:** <https://forum.microlist.org/t/issue-with-manager-for-ix83/2970>\
**Category:** Hardware\
**Tags:** micro-manager\
**Created:** [January 21, 2025, 2:41pm UTC](https://forum.microlist.org/t/issue-with-manager-for-ix83/2970 "2025-01-21T14:41:27Z")\
**Posts on this page:** 3\
**Page:** 1

<div class="post-metadata">

**Author:** ![lio14](https://avatars.discourse-cdn.com/v4/letter/l/a3d4f5/32.png) [@lio14](https://forum.microlist.org/u/lio14)\
**Post date:** [January 21, 2025, 2:41pm UTC](https://forum.microlist.org/t/issue-with-manager-for-ix83/2970/1 "2025-01-21T14:41:27Z")

</div>

I am trying to use the last µmanager version with the last pymmcore version on Windows 10 to control the microscope IX83 but I fall on the error message ‘Cannot find the adapter… mmgr\_dal…IX83’ whereas this adapter is present in the path…  
Does someone use µmanager with pymmcore and with which version both ?  
Thanks in advance,

Lionel

---

<div class="post-metadata">

**Author:** ![Hazen\_Babcock](https://sea2.discourse-cdn.com/flex020/user_avatar/forum.microlist.org/hazen_babcock/32/281_2.png) [@Hazen\_Babcock](https://forum.microlist.org/u/Hazen_Babcock)\
**Post date:** [January 22, 2025, 2:47pm UTC](https://forum.microlist.org/t/issue-with-manager-for-ix83/2970/2 "2025-01-22T14:47:22Z")

</div>

You tried the most recent nightly build of micro-manager? The latest release (July 13, 2021)? Or both and neither worked?

---

<div class="post-metadata">

**Author:** ![lio14](https://avatars.discourse-cdn.com/v4/letter/l/a3d4f5/32.png) [@lio14](https://forum.microlist.org/u/lio14)\
**Post date:** [January 22, 2025, 3:00pm UTC](https://forum.microlist.org/t/issue-with-manager-for-ix83/2970/3 "2025-01-22T15:00:38Z")

</div>

Hi Hazen,

I was using MM 11.1… instead of using a 10.1.1…

It is working now !

Lionel