# Coupling lasers to a widefield microscope?

**URL:** <https://forum.microlist.org/t/coupling-lasers-to-a-widefield-microscope/1979>\
**Category:** Hardware\
**Created:** [December 4, 2022, 7:57pm UTC](https://forum.microlist.org/t/coupling-lasers-to-a-widefield-microscope/1979 "2022-12-04T19:57:23Z")\
**Posts on this page:** 9\
**Page:** 1

<div class="post-metadata">

**Author:** ![Yossi\_Steinberger](https://sea2.discourse-cdn.com/flex020/user_avatar/forum.microlist.org/yossi_steinberger/32/1272_2.png) [@Yossi\_Steinberger](https://forum.microlist.org/u/Yossi_Steinberger)\
**Post date:** [December 4, 2022, 7:57pm UTC](https://forum.microlist.org/t/coupling-lasers-to-a-widefield-microscope/1979/1 "2022-12-04T19:57:23Z")

</div>

Dear μforum -

I have a Zeiss Axioobserver microscope, and a EMCCD camera, and wish to purchase a laser light source, and to set up a single molecule imaging system in my lab (I am a PhD student). However, I am ignorant of a good way to couple the lasers to the microscope.

In particular, I want to use the 100x, 1.4 numerical aperture objective, and to achieve a maximum power of 30 kW/cm^2 with a far-red laser of ~640nm, but to achieve much lower powers of 30 W/cm^2 with a 405nm, 488nm and 549nm laser. The illumination field should be at least 30 microns by 30 microns. I would like a Widefield, homogenous, square illumination field.

What do you recommend regarding laser engines and couplers, for a non-expert in optics, who is comfortable with reading the manual? A single component or very few components that I might purchase and easily assemble, in order to connect the high/low power lasers (uv,blue,red, and far-red), to the microscope, and to get Widefield illumination, would be ideal. Because the lab is an experimental biology lab, a coupling solution/part that couples the laser while completely encasing the laser beam is ideal.

I would be happy to hear also recommendations about which companies to consult with.

Thank you for your time -  
Yossi

Weizmann Institute of Science

---

<div class="post-metadata">

**Author:** ![Hazen\_Babcock](https://sea2.discourse-cdn.com/flex020/user_avatar/forum.microlist.org/hazen_babcock/32/281_2.png) [@Hazen\_Babcock](https://forum.microlist.org/u/Hazen_Babcock)\
**Post date:** [December 7, 2022, 2:25am UTC](https://forum.microlist.org/t/coupling-lasers-to-a-widefield-microscope/1979/2 "2022-12-07T02:25:06Z")

</div>

It sounds like you are interested in doing single molecule localization microscopy (SMLM)? If you haven’t already, probably the first person to ask is your Zeiss representative. They likely sell attachment(s) that would do what you are looking for and that would be easier to install and safe for most users.  
Lumencor ([https://lumencor.com/](https://lumencor.com/)) is one company that sells laser engines who might be worth contacting.

---

<div class="post-metadata">

**Author:** ![FishCsCells](https://sea2.discourse-cdn.com/flex020/user_avatar/forum.microlist.org/fishcscells/32/83_2.png) [@FishCsCells](https://forum.microlist.org/u/FishCsCells)\
**Post date:** [December 13, 2022, 9:20pm UTC](https://forum.microlist.org/t/coupling-lasers-to-a-widefield-microscope/1979/3 "2022-12-13T21:20:28Z")

</div>

I know the folks who created DNA PAINT had a super simple setup that included some laser pointers that they put together after the initial publication. Check out some of their papers–not the first, but something early.

---

<div class="post-metadata">

**Author:** ![pawlowska](https://sea2.discourse-cdn.com/flex020/user_avatar/forum.microlist.org/pawlowska/32/715_2.png) [@pawlowska](https://forum.microlist.org/u/pawlowska)\
**Post date:** [December 16, 2022, 6:01am UTC](https://forum.microlist.org/t/coupling-lasers-to-a-widefield-microscope/1979/4 "2022-12-16T06:01:25Z")

</div>

Purely from optics point of view:  
-if you illuminate the objective with a large collimated beam, you get a focused beam out  
-on the contrary, if you want to get a (roughly) collimated beam out, you need to focus your laser in the back focal plane of the objective. This means for instance placing a 20 cm lens something like 19-20 cm from the objective. The illumination will not be homogenous, but will rather have a roughly Gaussian profile.If you want to do better than that, you’ll have to use additional tricks.

Moreover, if you now have separate lasers, you’ll need a beam combiner.

---

<div class="post-metadata">

**Author:** ![Yossi\_Steinberger](https://sea2.discourse-cdn.com/flex020/user_avatar/forum.microlist.org/yossi_steinberger/32/1272_2.png) [@Yossi\_Steinberger](https://forum.microlist.org/u/Yossi_Steinberger)\
**Post date:** [June 12, 2025, 4:27pm UTC](https://forum.microlist.org/t/coupling-lasers-to-a-widefield-microscope/1979/5 "2025-06-12T16:27:51Z")

</div>

Dear μforum -

Thanks for the illuminating insights so far.

My question is still the same, but now more concrete.

**Technical summary of my desired setup**

- Microscope: **Nikon Eclipse Ti‑E** with **100×/1.4 NA** oil objective
- Laser: 637 nm OBIS‑FP, FC/APC, SM‑fiber (NA 0.09, core 4.5 µm)
- Goal: live‑cell HaloTag imaging (5 ms exposures) with a Gaussian spot whose central circular core is **30 µm in diameter** and maintains ≥ 87 % of peak intensity, yielding 3 kW cm⁻² within that region.

I will use, e.g., the Thorlabs Nikon Eclipse Ti Epi-Illumination Port Adapter (see image below) and lenses. After collimating the laser beam, and possibly also beam expanding it, do I need a lens to focus it onto the objective’s back focal plane (BFP), or does the Nikon Ti-E’s optical path inside the microscope body include lenses that perform this function?  
In a previous response, @pawlowska suggested placing a 20 cm lens 19–20 cm from the objective—does this imply I must provide the lens that focuses on the BFP, and how can I confirm the exact distance from the Epi-Illumination Port at which to place the lens?

I appreciate your time and expertise.

Warmly,  
Yossi

 ![image](https://us1.discourse-cdn.com/flex020/uploads/microlist/original/2X/5/50ed2e59187f8cdf0ce88b40101ed0322e0ecfc4.png)

---

<div class="post-metadata">

**Author:** ![Hazen\_Babcock](https://sea2.discourse-cdn.com/flex020/user_avatar/forum.microlist.org/hazen_babcock/32/281_2.png) [@Hazen\_Babcock](https://forum.microlist.org/u/Hazen_Babcock)\
**Post date:** [June 14, 2025, 1:45am UTC](https://forum.microlist.org/t/coupling-lasers-to-a-widefield-microscope/1979/6 "2025-06-14T01:45:54Z")

</div>

A Nikon Ti-E usually will not have a lens in the optical path between the filter and epi-illumination port. You could verify this by removing an objective and looking into the turret. If you can see an undistorted reflection of something like a cell phone light off of a dichroic in the filter wheel there is no lens in the optical path.

A 20cm lens could be a good choice, it depends on the diameter of your expanded and collimated input beam. The lens you choose and the objective will make a telescope with a demagnification proportional to the ratio of the focal lengths of the lens and the objective. If you use a 20cm lens and a 100x objective (approximately 2mm focal length for a Nikon) your expected beam size after the objective in the FOV is (input diameter) x 2mm/200mm. So if your input beam had a diameter of 10mm it would be 100um in the FOV. I think most people find it easier to use a longer focal length lens (so it is outside the body of the microscope) and a correspondingly large input beam diameter. Having the lens outside of the body also makes it easier to position.

How do you know you have the lens at the right distance? In my experience the easiest approach is to have the lens mounted on a rail or in optical cage so you can slide it back and forth. When the lens is at the right position the beam coming out of the objective is collimated. In practice you just adjust the lens position until the beam coming out of the objective is as small as possible at some point a good distance away from the microscope, like the ceiling of the lab. Then there is the X/Y alignment, for this it’s helpful to have the lens mounted on something like a Thorlabs CXY1A.

---

<div class="post-metadata">

**Author:** ![Hazen\_Babcock](https://sea2.discourse-cdn.com/flex020/user_avatar/forum.microlist.org/hazen_babcock/32/281_2.png) [@Hazen\_Babcock](https://forum.microlist.org/u/Hazen_Babcock)\
**Post date:** [June 14, 2025, 1:50am UTC](https://forum.microlist.org/t/coupling-lasers-to-a-widefield-microscope/1979/7 "2025-06-14T01:50:45Z")

</div>

This thread has some discussion of how to align a laser into a microscope:

> [@Basic microscopy Questions about building](https://forum.microlist.org/t/basic-microscopy-questions-about-building/2431):
>
> Hello everyone, Ive recently been tasked with building an inverted objective-tirf microscope. Im rather new to this so im learning by doing. I have built the system and aligned the laser, as far as I know. After aligning everything and connecting the cameras and hardware, I decided to do some test images. However, these images are a little strange. Perhaps you could help me out. The images were taken in the epifluorescence position (e. g the laser shining directly through the sample) with just …

---

<div class="post-metadata">

**Author:** ![Yossi\_Steinberger](https://sea2.discourse-cdn.com/flex020/user_avatar/forum.microlist.org/yossi_steinberger/32/1272_2.png) [@Yossi\_Steinberger](https://forum.microlist.org/u/Yossi_Steinberger)\
**Post date:** [July 11, 2025, 2:33pm UTC](https://forum.microlist.org/t/coupling-lasers-to-a-widefield-microscope/1979/8 "2025-07-11T14:33:31Z")

</div>

Dear Hazen and μforum,

Thank you for your guidance. I’ve designed a laser coupling system for our widefield microscope and would appreciate your feedback.

**System Overview:**

- **Source** : Fiber-coupled lasers (488nm/25mW, 561nm/200mW, 640nm/500mW) via a common single-mode polarization maintaining FC/APC fiber (Mean Field Diameter: 4.6μm @ 630nm)
- **Collimator** : 89mm focal length → 10mm beam diameter (1/e²)
- **Focusing lens** : 200mm focal length → 10μm diameter spot at back focal plane
- **Objective** : Nikon 100× objective → 174μm diameter illumination field

**Mechanical Design:** The collimator connects to the microscope’s back port via four precision rods. The focusing lens slides along these rods for alignment (see image).

 ![image](https://us1.discourse-cdn.com/flex020/uploads/microlist/original/2X/a/a947db052d4daad61f5dc2f69b36a79abbbf289d.jpeg)

**Questions:**

1. Does this optical design look correct for epi-illumination?
2. Will the four-rod mounting provide reliable alignment stability?

Thank you for any insights you can share.

Warmly,  
Yossi

---

<div class="post-metadata">

**Author:** ![Hazen\_Babcock](https://sea2.discourse-cdn.com/flex020/user_avatar/forum.microlist.org/hazen_babcock/32/281_2.png) [@Hazen\_Babcock](https://forum.microlist.org/u/Hazen_Babcock)\
**Post date:** [July 12, 2025, 2:47am UTC](https://forum.microlist.org/t/coupling-lasers-to-a-widefield-microscope/1979/9 "2025-07-12T02:47:58Z")

</div>

Well my calculation is that you will end up with a 100um diameter illumination field given a 10mm beam from the collimator. You have a 200mm focal length lens and a 100x objective has a nominal focal length of 2mm (Nikon → 200mm tube lens → 100x mag → 2mm focal length) so you have a telescope that will reduce the beam size by a factor of 100 (2mm/200mm). That said, calculation only gets you so far, and sometimes it is better to just build it and find out whether the beam diameter is workable or not, assuming you have the time and resources to iterate if it is not workable.

Does the 2" 200mm lens fit through the hole in the microscope’s back port? My recollection is that the back port is more than 200mm from the back focal plane of the objective so a 200mm lens would have to be inside the frame of the microscope. For this application a 1" 200mm lens is probably fine and I don’t think you will have issues with beam distortion.

You probably want to support the outer end of the ER rods so that they are less sensitive to vibration. If the microscope is on a optical bread board you could add a post that comes up from the bread board. If it isn’t a post that just sits on the table the microscope is on is probably sufficient.