# Autofocus in Z for MDA not changing z range

**URL:** <https://forum.microlist.org/t/autofocus-in-z-for-mda-not-changing-z-range/2247>\
**Category:** Software\
**Tags:** micro-manager, confocal\
**Created:** [August 17, 2023, 4:39pm UTC](https://forum.microlist.org/t/autofocus-in-z-for-mda-not-changing-z-range/2247 "2023-08-17T16:39:48Z")\
**Posts on this page:** 2\
**Page:** 1

<div class="post-metadata">

**Author:** ![sommer.a](https://avatars.discourse-cdn.com/v4/letter/s/df788c/32.png) [@sommer.a](https://forum.microlist.org/u/sommer.a)\
**Post date:** [August 17, 2023, 4:39pm UTC](https://forum.microlist.org/t/autofocus-in-z-for-mda-not-changing-z-range/2247/1 "2023-08-17T16:39:48Z")

</div>

I am trying to use autofocus to focus in the z direction for an MDA acquisition (multiple timepoints, channels, slices). Looking at the code for OughtaFocus, it appears that the top z slice +/- the user-given search range is used. Then, based on the sharpest image, the z position is changed. However, for the next acquisition, that new z value is not used for the top, middle, or bottom of the stack. The z position in the metadata is the same set of values for top/bottom/step for all time points. It seems that after autofocus, the z moves back to the top of the stack for the next acquisition. In the Acquisition engine, lines 232-243 (linked below), autofocus is run and there is a hook to return to the top of the z range for the next timepoint.

**Main question:** Why is the z position found from autofocus not used to adjust the z range in an MDA acquisition, and how can this be done?

I can change the code for the oughtafocus plugin-- I have changed the search range from top z +/- half the search range to middle z +/- half the search range, and what I want to do is use the newly focused z as the middle of the z stack at the next time point, and use that as the middle of the search range for the next round of autofocus. This seems intuitive to me, and I want to implement it without interfering with the acquisition.

> <https://github.com/micro-manager/micro-manager/blob/28dccce3ad96ee62395afb48cd82b679e8b00014/mmstudio/src/main/java/org/micromanager/acquisition/internal/acqengjcompat/multimda/acqengj/MultiAcqEngJAdapter.java#L90>

---

<div class="post-metadata">

**Author:** ![nico.stuurman](https://sea2.discourse-cdn.com/flex020/user_avatar/forum.microlist.org/nico.stuurman/32/302_2.png) [@nico.stuurman](https://forum.microlist.org/u/nico.stuurman)\
**Post date:** [January 10, 2024, 1:28am UTC](https://forum.microlist.org/t/autofocus-in-z-for-mda-not-changing-z-range/2247/2 "2024-01-10T01:28:43Z")

</div>

For your Z-stack, do you use absolute or relative positioning? Autofocus will not work with absolute positioning, but it will work with relative (it should set the “zero” of the relative position to the new in focus Z position).