# 96-well plates you would recommend for high NA imaging

**URL:** <https://forum.microlist.org/t/96-well-plates-you-would-recommend-for-high-na-imaging/1868>\
**Category:** Hardware\
**Created:** [July 25, 2022, 10:55am UTC](https://forum.microlist.org/t/96-well-plates-you-would-recommend-for-high-na-imaging/1868 "2022-07-25T10:55:18Z")\
**Posts on this page:** 8\
**Page:** 1

<div class="post-metadata">

**Author:** ![NicoKiaru](https://sea2.discourse-cdn.com/flex020/user_avatar/forum.microlist.org/nicokiaru/32/451_2.png) [@NicoKiaru](https://forum.microlist.org/u/NicoKiaru)\
**Post date:** [July 25, 2022, 10:55am UTC](https://forum.microlist.org/t/96-well-plates-you-would-recommend-for-high-na-imaging/1868/1 "2022-07-25T10:55:19Z")

</div>

Hello everybody,

I’d like to know if you have some recommendations for 96-well plates (well not necessarily 96) that would be suitable for high-NA imaging.

Ideally, the well plate would:

- have a very homogeneous thickness, of the order 170 - 190 um
- have no skirt, thus allowing to focus all wells with high-NA objectives
- be reasonably cheap

The well-plates we recommend at our facility is the [PhenoPlate 96](https://www.perkinelmer.com/uk/product/cellcarrier-96-ultra-lid-8x20b-6055302) from Perkin-Elmer, but we would like to know alternative options.

Thanks!

Nicolas

(this will be crossposted on the confocal list)

---

<div class="post-metadata">

**Author:** ![manerotoni](https://sea2.discourse-cdn.com/flex020/user_avatar/forum.microlist.org/manerotoni/32/398_2.png) [@manerotoni](https://forum.microlist.org/u/manerotoni)\
**Post date:** [July 29, 2022, 8:44am UTC](https://forum.microlist.org/t/96-well-plates-you-would-recommend-for-high-na-imaging/1868/2 "2022-07-29T08:44:36Z")

</div>

Hello,  
we use often those [  
https://zell-kontakt.de/products/imaging-plate/imaging-plate-glasscover-2/](https://zell-kontakt.de/products/imaging-plate/imaging-plate-glasscover-2/)

They are fine. I normally avoid to image the outer wells as it get quite tight. It depends on how bulky your objective is.

Antonio

---

<div class="post-metadata">

**Author:** ![JamesOrth](https://sea2.discourse-cdn.com/flex020/user_avatar/forum.microlist.org/jamesorth/32/153_2.png) [@JamesOrth](https://forum.microlist.org/u/JamesOrth)\
**Post date:** [August 8, 2022, 8:30pm UTC](https://forum.microlist.org/t/96-well-plates-you-would-recommend-for-high-na-imaging/1868/3 "2022-08-08T20:30:46Z")

</div>

You may have had other replies, we’ve had good luck with 96 well plates from CellVis, they use 1 piece of #1.5H German glass and are well priced. They do have a skirt so you likely can’t image the entire plate with high-NA low WD lenses.

---

<div class="post-metadata">

**Author:** ![phoTONIcs](https://avatars.discourse-cdn.com/v4/letter/p/45deac/32.png) [@phoTONIcs](https://forum.microlist.org/u/phoTONIcs)\
**Post date:** [August 8, 2022, 10:21pm UTC](https://forum.microlist.org/t/96-well-plates-you-would-recommend-for-high-na-imaging/1868/4 "2022-08-08T22:21:46Z")

</div>

[https://www.auroramicroplates.com/#products](https://www.auroramicroplates.com/#products)

---

<div class="post-metadata">

**Author:** ![NicoKiaru](https://sea2.discourse-cdn.com/flex020/user_avatar/forum.microlist.org/nicokiaru/32/451_2.png) [@NicoKiaru](https://forum.microlist.org/u/NicoKiaru)\
**Post date:** [August 23, 2022, 10:27am UTC](https://forum.microlist.org/t/96-well-plates-you-would-recommend-for-high-na-imaging/1868/5 "2022-08-23T10:27:34Z")

</div>

Hi everybody,

Thank you for your answers! I’ve made a google sheet with the suggestions coming from this forum and with the ones from the confocal list:

> **[A selection of 96 well plates for High NA imaging](https://docs.google.com/spreadsheets/d/1b-hA_khruxoGfJ4QfVnz98Wsb_EEOZrUn7gL7vUfm88/edit?usp=sharing)**
>
> Feuille 1
> 
> The well plates listed should be compatible with high-NA imaging, meaning:
> \- have a very homogeneous bottom thickness, of the order 170 - 190 um
> \- have no skirt, thus allowing to focus all wells with high-NA objectives
> Brand,Link (96...

If you want to edit it to add some more, please let me know and I’ll give you edit rights.

Nicolas

---

<div class="post-metadata">

**Author:** ![jdelgadozmb](https://sea2.discourse-cdn.com/flex020/user_avatar/forum.microlist.org/jdelgadozmb/32/1042_2.png) [@jdelgadozmb](https://forum.microlist.org/u/jdelgadozmb)\
**Post date:** [September 5, 2022, 8:57am UTC](https://forum.microlist.org/t/96-well-plates-you-would-recommend-for-high-na-imaging/1868/6 "2022-09-05T08:57:55Z")

</div>

Thanks for compiling the list! We also have good experience with  
Corning 4580 half area glass bottom plates if you’d like to add

---

<div class="post-metadata">

**Author:** ![NicoKiaru](https://sea2.discourse-cdn.com/flex020/user_avatar/forum.microlist.org/nicokiaru/32/451_2.png) [@NicoKiaru](https://forum.microlist.org/u/NicoKiaru)\
**Post date:** [September 5, 2022, 10:22am UTC](https://forum.microlist.org/t/96-well-plates-you-would-recommend-for-high-na-imaging/1868/7 "2022-09-05T10:22:02Z")

</div>

Thanks, I’ve added it, but I couldn’t find the skirt size unfortunately

---

<div class="post-metadata">

**Author:** ![jdelgadozmb](https://sea2.discourse-cdn.com/flex020/user_avatar/forum.microlist.org/jdelgadozmb/32/1042_2.png) [@jdelgadozmb](https://forum.microlist.org/u/jdelgadozmb)\
**Post date:** [September 6, 2022, 2:59pm UTC](https://forum.microlist.org/t/96-well-plates-you-would-recommend-for-high-na-imaging/1868/8 "2022-09-06T14:59:24Z")

</div>

Details can be found here

> **[CLS-AN-437\_DL.pdf](https://www.corning.com/catalog/cls/documents/faqs/CLS-AN-437_DL.pdf)**

Skirt Height (mm) 6.096

Here additional dimensions, unfortunately they do not provide a drawing

> **[MicroplateDimensions96-384-1536.pdf](https://www.corning.com/catalog/cls/documents/drawings/MicroplateDimensions96-384-1536.pdf)**